Varian 800 MIB Help!

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
So I have an HPLC problem, I have 2 210 pumps(A & B), a 800 MIB converter, and a 355 RI detector. The software I am using is galaxie configuration manager. My problem is that the software does not detect the 355 RI detector and only detects Pump A. Is it the way I am connecting the relay? The connection is as follows: PC->Converter->Pump A->Pump B.
Thanks!
Myca
Tìm dịch vụ giup viec nha uy tín - giúp việc nhà quận 6
Hello,

The 355 RI detector is not supported by Galaxie.
Only the 356-RI detector was supported, but not using the Galaxie Varian LC package, but only the Varian Modular LC package.
As Galaxie is no more supported anymore, and as the licenses are not sold since almost 2 years, I'm afraid you have no solution to control your RI using Galaxie...
You still can get the analog data using MIB, in this case, the detector will not be controlled.
you have to setup the MIB driver (800 analog when adding driver in the instrument), and to setup the right analog port.