Hello my friends,

I´m facing a problem with my GC Agilent 6850.
When I turn it on, the gas flow of the gases are:

Air: 477 (even when I set to 0). It doesn´t change.
H2: 11.6 (even when I set to 0). It doesn´t change.
Make-Up N2: it works as I set.

What should I do?

Thanks,

Lucas.
Lucas - if I had this issue, first thing I would do is use my accessory electronic flow checker or soap bubble meter and see if such flows are really as flow meter says.

Are you using H2 for both carrier gas and for the FID ?
I already did it. And there are no flows coming out.

No. I´m using H2 only for FID. For carrier gas I´m using N2.

Any idea?
Possibilities :

-Loose contact between EPC and EPC control board , check ribbon cable connections
-Faulty EPC
-Faulty EPC control board