Hi
I'm using a HP 5890gc/5971ms with chemstation and get an error after I click start run.
Run Error: -302,"(:RUN:RDY ON;): Timeout Waiting 1286000 usec for DP to process 1 scan(s).",-302,"(RUN:RDY ON;): Error Executing

Any idea what's going on?
Looks as if the instrument is entering the ready state then waiting approximately 20minutes for a start signal that never arrives.

What is your setup? Are you using an autosampler or other sample introduction device? If no autosampler do you press "RUN" at the time of injection?

Also you have to have the Remote cable connected from the back of the MS to a connector on the top of the right side of the GC, usually to a small rectangular connector mounted under the cover on the top right side above and behind the display. If that cable is not connected the MS will not receive the start signal and give a timeout error.