by
Earl C » Mon Jan 04, 2010 10:27 pm
The difference in the two valve part numbers should not have been a problem, but I dont know that we can rule that out entirely. Trying another one couldn't hurt but could be a waste of time.
Just curious if you have checked the settings on the main EPC board? See below for instructions from agilent site. The board switches are on the top right corner, red housing in your pics. If I remember correctly for split/splitless they are all to the right. Kind of looks right in your pic but can't really tell. What is your zero pressure reading? Set your injection port pressure to 0 then press gold key then 2 for port A, 3 for port B. If the zero reading is higher than the set point you would not have pressure control.
BTW, love the Prof. Frink on your desk!
Copied from:
http://www.chem.agilent.com/Library/Sup ... a16099.pdf
Prior to troubleshooting a problem with the flow system of the GC it is prudent to check first the EPC control board configuration and the Zero on the EPC Channel.
NOTE: The EPC board is located on the right hand side of the GC within the main panel and it is plugged into the mainboard of the GC. If at this point you are unsure about how to check these switch settings, please call Agilent. DO NOT PROCEED.
The switch settings for the 5890 Series II EPC:
IN A1 OR IN B1 Right, currently unused PID
IN A0 OR IN B0 Left, Programmable Cool On Column (PID)/ Right, Purged Packed Inlet/ Right, (BPR) Split/Splitless Capillary Inlet
MODE A OR MODE B Left, (FPR) Programmable Cool On Column Inlet & Purged Packed Inlet /Right, (BPR) Split/Splitless Capillary Inlet
PID= Proportional/ Integral/ Differential heated zone
FPR= Forware Pressure Regulating mode
BPR= Back Pressure Regulating mode
Zeroing the EPC channel is necessary to correct for tolerances of the sensor and other electronic components. Zeroing the channel nullifies any offset that would keep a “0â€