Hi, does anybody have knowledge/experience of using OpenLab CDS to create a custom calculation. I want to compute the %Recovery of a system check sample against the reported/quantitated amount obtained for a calibration standard.
I was thinking this would be quite straightforward however i am failing miserably and would much appreciate any help offered.
Thanks
Hi ,

Been a while since I have used Open lab

% Recovery should not need a custom field if you add weights and dilutors /multipliers to sample and standards v calib curve (area v amount for stds)

reference worth looking at :

https://www.agilent.com/cs/library/user ... sGuide.pdf