Agilent OpenLab2 Custom calculations

Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.

2 posts Page 1 of 1
Hi, does anybody have knowledge/experience of using OpenLab CDS to create a custom calculation. I want to compute the %Recovery of a system check sample against the reported/quantitated amount obtained for a calibration standard.
I was thinking this would be quite straightforward however i am failing miserably and would much appreciate any help offered.
Thanks
Hi ,

Been a while since I have used Open lab

% Recovery should not need a custom field if you add weights and dilutors /multipliers to sample and standards v calib curve (area v amount for stds)

reference worth looking at :

https://www.agilent.com/cs/library/user ... sGuide.pdf
2 posts Page 1 of 1

Who is online

In total there are 15 users online :: 0 registered, 0 hidden and 15 guests (based on users active over the past 5 minutes)
Most users ever online was 234 on Tue Feb 06, 2018 7:33 am

Users browsing this forum: No registered users and 15 guests