Howdy everyone! First time posting here, long time lurker as a non-registered member.

So I'm a GC-MS method (a method that honestly could be on a FID) and I was given my samples to analyze on this method. I prepared my ISTD (lot 1) for this method and I had more "last minute samples" come in and I didn't account for the additional samples when preparing my initial ISTD batch. So these samples are the "worlds supply" and have been diluted with my ISTD (lot 1) diluent. I prepared another batch of ISTD (lot 2) for my 5 pt calibration curve at the same concentration as lot 1.

So here's my possible issue, lot 1 conc is 3.02 mg/mL and lot 2 is 2.99 mg/mL. I haven't started the sequence yet because I realized there might be an issue if I were to calculate the samples true concentration as the RF might be different.

Would it be wise to use relative RF or Area % to calculate my "worlds supply" of samples w/w%?

The problem is the method is pretty old (contains chart speed parameters, lol) and the person who developed it and validated the method isn't available and there's no commentary in the method about relative RF. I just want to ensure I give out as accurate results as possible on this. Any assistance is greatly appreciated!
For 1% difference in internal standard concentration?
2 posts Page 1 of 1

Who is online

In total there are 37 users online :: 1 registered, 0 hidden and 36 guests (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: Google [Bot] and 36 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry