I was asked to create a new method on a GC (New to that process). I got the method set up approved and entered samples into the sequence table. I went to run the sequence (unit said it was ready) and as soon as I hit run, it attempts to run it and then it just says sequence stopping without running anything. I have no idea what is wrong. Please advise.
-
- tom jupille
- Site Admin
-
- Posts: 4954
- Joined: Wed Aug 11, 2004 4:55 pm
- Location: Alamo, CA / USA
It would help a lot if you mentioned what hardware and software you are using.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 293
- Joined: Mon Sep 13, 2010 8:28 am
I'd rather say: you have sequence line which loads non-existent or broken (with zero analysis time) instrument method.
-
- Posts: 5260
- Joined: Thu Oct 13, 2005 2:29 pm
- Location: Maun Botswana
Welcome to the forum.
My advice is, post the details of what you are doing, on what instrument, with what method. With the minimum description of the problem you posted, none of us has an idea what is wrong either.
Rod's law: the more you don't tell us, the more we can't help you.
Peter
My advice is, post the details of what you are doing, on what instrument, with what method. With the minimum description of the problem you posted, none of us has an idea what is wrong either.
Rod's law: the more you don't tell us, the more we can't help you.
Peter
Peter Apps
-
- Posts: 5
- Joined: Fri Sep 11, 2020 8:00 pm
I love your username and I can relate so much! I now have two years experience and I can promise it gets better.
Were you able to resolve your issue?
Were you able to resolve your issue?
5 posts
Page 1 of 1
Who is online
In total there are 26 users online :: 0 registered, 0 hidden and 26 guests (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 26 guests
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 26 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science, @Food_Science_, @Pharma_Science_
- Follow us on Linkedin: https://www.linkedin.com/showcase/separation-science-/