Are gas filters an issue when switching between He & N2 carrier gas? The following Agilent installation instructions indicate a long purge may be necessary to remove the nitrogen.

"Flush the filter. Set the instrument split vent flow (purge flow) to 400 mL/min, then:
• If using a Moisture Filter in combination with an MS detector, purge the filter for at least 60 minutes. This will flush the filter of dry nitrogen so that it does not appear as a background ion in your MS analysis."
markelh wrote:
Are gas filters an issue when switching between He & N2 carrier gas? The following Agilent installation instructions indicate a long purge may be necessary to remove the nitrogen.

"Flush the filter. Set the instrument split vent flow (purge flow) to 400 mL/min, then:
• If using a Moisture Filter in combination with an MS detector, purge the filter for at least 60 minutes. This will flush the filter of dry nitrogen so that it does not appear as a background ion in your MS analysis."

Install separate filters for helium and nitrogen. Otherwise you will have mentioned problem.
Agreed, use the filters before the switching valve and you only need to purge for a few minutes to remove the nitrogen from the actual lines and injection port.

I usually let mine run for 15-20 minutes at a more normal 20ml/min split flow before starting analysis and it seems to work.
The past is there to guide us into the future, not to dwell in.
3 posts Page 1 of 1

Who is online

In total there are 20 users online :: 0 registered, 0 hidden and 20 guests (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 20 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry