Running an Agilent GC/MS system (6890/5975, MSD Chemstation with CTC control add-on) with GC-PAL autosampler. Been running for years with no major issues. Last week, the helium ran out and so the sampler was stuck in the "waiting for ready". Changed cylinder, reset everything and now the CTC gets the command to start the injection, waits for system ready, injects the sample, sends the remote signal to the GC/MS to start the run but then... it just stays there at the GC inlet instead of going on with its usual cycle of cleaning and going home until the next injection? CTC Status message in MSD Chemstation just stays at "RUNNING - Injecting into GC Inj1". AFter the method has finished running, the CTC then no longer responds (if I try to reset the injector through MSD Chemstation, it's saying it reset it but it clearly did not, or the software just freezes). Now, if I try to run a method outside of the software through the CTC control pad, it runs as expected.

I have tried pretty much everything I can think of! Full cold boots, changing the COM port, different method, configuring a new GC/MS instrument in the MSD software, different cables for both the serial/USB connection and the remote cable, running with GC only configuration (no MS), resetting defaults in the injection method, selecting a different injection method from those available, reflashing the CTC with an older backup .sss file, as well as removing and re-installing the Agilent control for CTC software... NOTHING WORKS.

If anyone has come across a similar issue or has ANY idea of what I should be looking at, it would be greatly appreciated as I'm at a complete loss after several hours.
Just to follow-up in case someone comes across a similar issue. After pretty much eliminating every other possibility (including dragging out the old XP computer that used to run it), we've established it was the board inside the CTC itself (confirmed by switching with another one).
2 posts Page 1 of 1

Who is online

In total there are 17 users online :: 0 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 17 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry