front tailing
Page 1 of 1
Which factors can cause a front tailing in HPLC?
for some basic compounds it can be overloading of residual silanols at certain pH
Vlad Orlovsky
HELIX Chromatography
My opinions might be bias, but I have about 1000 examples to support them. Check our website for new science and applications
www.helixchrom.com
www.helixchrom.com
Thnks Vlad,
But the analyte does not contain nor basic neither acid groups.
Inject less microliters, see if it gets better.
reasons:
overloading
poorly packed column (with fronting to help peaks with tailing)
diluent effect
solubility effect
tautomerism of compound (not defined "batman peak" for example)
overloading
diluent effect
solubility effect
Vlad Orlovsky
www.helixchrom.com
www.helixchrom.com
- tom jupille
- Site Admin
There is a chalk-talk that covers that:
viewtopic.php?f=31&t=18426
Fronting is discussed near the end (about 8 minutes in), but it's best to watch it all to put fronting into perspective.
Page 1 of 1
In total there are 9 users online :: 2 registered, 0 hidden and 7 guests (based on users active over the past 5 minutes)
Users browsing this forum: Ahrefs [Bot], Bing [Bot] and 7 guests
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science