RP Column Stability Under Reducing Conditions
2 posts Page 1 of 1
I'm trying to analyze some compounds that must be reduced (and stay reduced) to be water soluble by UPLC. I am adapting a method our lab previously used with CE where we included 20 mM thiourea dioxide (Eo ~ -1000 mV, yikes) in the mobile phase buffer at pH 9.3.
My question is whether or not thiourea dioxide (or any other comparably strong reducing agent) is going to damage my C18 column. It is a Waters Acquity BEH C18 UPLC column. I've asked Waters and they didn't have an answer. Does anyone have any experience doing RP LC under strong reducing conditions? The rest of the method is typical, involving MeOH/ACN, ammonium acetate buffer, column at 25 C / 40 C, etc.
Thanks a bunch!
This paper might help you, just a little google!
Good luck
Oxidation and Decomposition Kinetics of Thiourea Oxides http ...
www.paper.edu.cn/scholar/.../gaoqingyu-20070... - Diese Seite übersetzen
In this report the decomposition and oxidation kinetics of thiourea dioxide and thiourea ... ion-pair high-performance liquid chromatography (HPLC) method.
Gerhard Kratz, Kratz_Gerhard@web.de
2 posts Page 1 of 1
In total there are 9 users online :: 0 registered, 0 hidden and 9 guests (based on users active over the past 5 minutes)