Hello All,
I have a student who wishes to conduct an analysis of fatty acids. They are Anandamide (arachidonylethanolamide), 2-AG (2-Arachidonylglycerol), and Palmitoylethanolamide. Does anyone have any ideas about the required conditions / mobile phase (I can only do isocratic analysis) or where I should look?
thanks
Jon
Jon,
I happen to have some experience analyzing surfactant related compounds. I believe these three are in this catagory. Here is what I would start:
Column: a RP column (C8 or C18), 4.6 x 150 mm, 5 micron should be sufficient
Mobile phase: acetonitrile/D.I. H2O (or NH4OAc buffer for better ELSD response)
Flow rate: 1 mL/min
Detection: ELSD
Here is a link for mor information on surfactant analysis using Acclaim Surfactant column, for your information.
http://www1.dionex.com/en-us/columns_ac ... 25924.html
Good luck,
XL
Thx XL, that's a good place for us to start
To me, these are just common neutral organics,