analysis of polyacrilic acid in ophthalmic solution
have some method of analysis of polyacrylic acid in ophthalmic solution?
HPLC or GC method?
HPLC or GC method?
Too big for GC.
HPLC or GC method?
*just get you a simple gel column with H+ cation exchange capability and use USP as the mobile phase on an LC.
Thank you.
I have some questions:
-such as the treatment of the sample?
-not be filtered with 0.45um filter pore size when diluted with water.
have you any application or description of the method?
You can use Primesep C column in anion-exclusion mode similar to algininc acid, which is also an acidic polymer:
http://www.sielc.com/Compound-Alginic-Acid.html
This method is validated in two companies.
Vlad Orlovsky
HELIX Chromatography
www.helixchrom.com
Vlad's suggestion looks fantastic based on the link provided.
If you have an SEC column handy you may want to try that too since your large polymer molecule will elute first, followed by smaller molecules. if you have other UV active large molecules in your solution you will need to make sure they don't co-elute.
MestizoJoe
Analytical Chemist and Adventurer
Venture Industries
Spider-Skull Island
here is an article on analysis of aginic acid in Oral suspension using Primesep C column. I know at least two other places which use similar method
http://sphinxsai.com/s_v2_n2/PT_V.2No.2 ... -1638).pdf
Acrylic acid has pKa close to alginic acid. Algininc acid is a polymer with the wide range of MW, but it still elutes as a relatively narrow peak.
Vlad Orlovsky
www.helixchrom.com