Anion exchange column needed
Dear all,
I am looking for a column that fits the European Pharmacopeia specifications of one packed with an "anion exchange resin R1". It needs to be 7um and in 150 x 4.6mm dimensions.
I have found a PRP-X100 and a MCI Gel SCA04 but these are only available in 5um particles sizes.
Does anyone know of a different column that fits the above requirements?
Thanks!
Sball,
You are correct, the PRP-X100 is only available in 3, 5 and 10µm particle sizes. The PRP-X110 however is available in 7µm particle size (4.6x150mm). The capacity of the X110 is lower compared to the X100 but otherwise the columns are identical. And I apologize for the rather product specific posting.
Best regards,
Robert Haefele
Hamilton Company
rhaefele@hamiltoncompany.com
Robert, there is no reason to apologize. You had the expertise that could answer the question. Very few other people would have known that fact.
Thanks a lot for your help Robert. I hadn't heard of that column before.