Do you have the option of temporarily running your MS from a higher-purity N2 cylinder, to see if that makes a difference?
any ideas how the N2 generator could be the source of contamination. Wouldn't you need a solvent to come into contact with N2 parts to leach the plasticizers out? The only liquid is water vapor from the air.
Tedinelkgrove,
I'm afraid I have no other N2 to switch to but the one from the N2 generator. I will try to mount the trap again in the system, though. My problems seem to have started somewhere around the time it was removed, so it might be a source of at least some problems.
Dejan Orcic
Asst. prof.
Department of Chemistry, Biochemistry and Environmental Protection
Faculty of Sciences, Novi Sad, Serbia
Asst. prof.
Tedinelkgrove,
What happened after you reinstalled the trap? Did the 214 disappear? I am having the same issue with my instrument, I have a LTQ Orbitrap XL. The 214 appeared about a year ago and we have tried everything we could think of for troubleshooting but the 214 is always present.
Celine
Page 2 of 2