PAHs by LC-MS/MS
Hi All,
Is it possible/advantageous to use LC-MS/MS for the analysis of PAHs?
Is there any standard method using LC-MS/MS for PAHs?
Thanks in advance!
Possible ?---yes (MAYBE). Advantageous ?---NO.
PAH's analyzed by LC/MS (ESI or APCI) give very stable [M+H]+ ions---difficult to fragment, especially to differentiate the isomers.
You might research APPI (atmospheric pressure photo-ionization) to see if that is more promising than ESI/APCI.
However, why re-invent the wheel ?---there are well-established EI/GC/MS methods, which have the advantage of very high plate counts of capillary columns to resolve the isomeric PAH's. There is also the matter of identification from an EI spectrum; there are commercial databases (NIST etc) for library search that are sold with GC/MS systems.
Thanks for your reply.
Actually, our GC-MS system is not behaving well with high molecular weight PAHs!! :wink:
If it is loss of linearity have you tried the larger draw out plate for the ion source?(Assuming it is an Agilent)
Thanks.
It is an Ion Trap GC-MS. I get dirty chromatogram at temperatures around 300 oC and above.
