Altering gradient to elute sample at a later retention time
I am using the Waters H-class UPLC with an ELS detector and the acquity c18 beh column. Based on my current gradient, the sample I want to analyze is eluting at a retention time of 0.4 minutes. Could someone please guide me on how to try and alter my gradient such that I can get my sample to elute at a later retention time?
The current gradient I am using is
Time(mins)/ Flow rate(mL/min)/ aqueous%/ organic%
0/ 0.32/ 60/ 40
3.33/ 0.32/ 70/ 30
6.66/ 0.32/ 80/ 20
10/ 0.32/ 90/ 10
12/ 0.32/ 60/ 40
Thanks!
Did you write that correctly? Are you actually *decreasing* the organic during the gradient??
I would try a HILIC column, and if that will not give good retention time an IEX separation. Good luck.
What is the nature of your analyte? Neutral, acidic,basic or zwitter-ionic?
Vlad Orlovsky
HELIX Chromatography
www.helixchrom.com
Users browsing this forum: Google [Bot] and 9 guests