Electron multiplier, change, sensitivity
I couldn't find a simple answer to it really. When electron multiplier should be changed?
The detector and EDR calibration report shows optimum of the detector in my instrument of 1700 V (Varian 1200L MS/MS instrument). For example in March last year it was 1600 V. I suppose it's not such a big difference but in Varian tune report there's another voltage value of which autotunning is proceeded with. Now it's 1700 V but last March it was 1100 V. I actually don't understand why this number was lower than the optimum.
Can the change of electron multiplier help towards better sensitivity?
Thank you for any help.
Thank you,
Kamila
electron multipliers become less sensitive with use. Their gain depends on the voltage applied to them. To compensate for the reduced sensitivity with use, the instrument will continually increase the voltage to maintain the same gain. The optimum voltage isn't a constant; it is measured during tuning, and gradually increases as the electron multiplier ages.
The reason there is an optimum is that if you increase the voltage further, there will be increased electronic noise, and increased risk of the electron multiplier failing completely (at some voltage, it will fail).
Different instruments and different manufacturers will have different target gains, so don't be surprised if apparently the same electron multiplier works at different voltages in different instruments. This also affects their average life-span in the instrument concerned.
So your question: when to change them? It depends how tolerant you are of failures. You should certainly change them when they reach a voltage that the instrument says is the maximum. You may find that methods are getting unacceptably noisy as they reach higher voltages; if so, note the voltage, and change them when (or just before) they reach this stage. If you find that they're failing altogether, again, consider setting up a regime where you change them 100V before the voltage they reached before sudden failure.
Unless your electron multiplier is very old, and probably maxed-out on voltage, changing it is an expensive option to increase sensitivity; for an average electron multiplier half-way through its life, it might, slightly, or it mightn't at all.