Help regarding baseline problem
I have problems in base line setting with HPLC software.
The baseline is not proper and gets shifted when a sample or standard is injected. Also, we are not able to get area of the peak from the chromatogram.
I have attached pictures of the chromatogram.
Please help in rectifying these problems and carryout the processes in place.
Instrument details:
Shimadzu UFLC20AD
Model: CTO-10ASVP
RID: RIDLIOA
Diode detector: SPD M20A
Software: LCsolution software
Expecting a reply at the earliest.
Do you experiance the same baseline shape when you inject a blank sample i.e. solvent? If you havn't tried it, that would be an important action. As well as injecting nothing - i.e. zero injection.
when injecting blank, the baseline is straight. Zero injection is also fine.
But about 5 minutes after injecting the sample or standard (after the 1st peak) there is unsteadiness in baseline as you see in the picture.
I am really conerned about this and donot know why its happening.
help me in sorting this out.
OK. That means the samples contain something (s) that elutes in an uncontrolable manner i.e. doesn't form a peak or peaks. That something absorbs ligth at the working wavelemght. The phenomenon is often observed when ionized substances elute in pH uncontrolled mobile phase.
My suggestion: lower the mobile phase pH to something like 2.5 using a buffer. Could be phosphate. Or add some TFA to your current mobile phase.
Best regards
Hi sathyavrathan
What is the mobile phase? In what solvent you dissolve your sample?
Best regards
Fernando
Thanks danko.
I am using 5mM H2SO4 as mobile phase and HPLC water to dissolve the sample.
Its all done as given in the literature.
what else may the cause for the problem??
OK – not a buffer, but might be OK.
Please let me know:
1. What does the sample matrix consist of (some salt/buffer, or some polymer?)
2. What is the pH in the sample solution
3. What is the mobile phase organic modifier Acetonitrile, Methanol, Propanol….
4. Detector wavelength choice
Then we might find a solution to your problem.