6890 Pneumatics
I am having inlet pneumatics problems with a Agilent 6890A. It cannot stabilize the pressure. I have replaced the pneumatics board and have tried a new pneumatics module, but it did not help, so I put the old one back in. It will pass set point by a couple of psi then drop below a couple of psi, but does not ever stabilize. Does anyone have any ideas?
I may have found the problem, but am not sure. Went into diagnostics/back inlet/pneumatics module and the temperature is incorrect (shows 80-84C and is not stable). All other instruments are ~35C and pretty stable. Must have gotten a bad pneumatics controller board.
I had that once with a Tekmar 7000 headspace sampler hooked up. If pumbed the way they suggest the inlet feed goes through an additional resistance frit for manual inlet control that caused issues. I replumbed it bypassing the frit and there were no more issues. It sounds like you may have a plug in the lines somewhere. One place to check is split vent tubing. That stuff fills with condensed tar (it is unheated) and can become blocked even.
Users browsing this forum: Google [Bot] and 5 guests
