How to Integrate 35900E signal in Chemstation C.01.06
Hi all,
I have existing 1200 chemstation instrument and is connected to the 35900E Series II ADC.
There is no problem to interface both instrument to the PC. I can see the online plot from Chemstation C.01.06.
The problem that I faced now is couldn't capture the signal in the windows when i run the sequence.
The window is like when we run a sequence, the signal will be captured and we can integrate the data. However, it couldn't capture the 35900E signal.
Kindly advise is there any setting that I missed out?
I've run across this exact issue before: signal shows in Online Plot but is not saved in the run or sequence (no vertical red line at injection in online plot).
You need a remote cable from the 35900 box to the stack, like on ours the 35900 remote cable connects to remote on the UV detector. Got it?
Thanks for the advice. It is work once connected the remote cable from 35900E to the 1200 stack. It really made my day. Thanks for the sharing.
Winner, winner, chicken dinner !
Part of being smart is remembering and learning from my past mistakes...
Is he going to E-mail that chicken dinner to you?