I just cleaned the source, changed the column of my GC MS but I see a strange contamination (see attached file)
Any idea about what this could be?
Thanks a lot,
Kevin
m/z 191 & 207 are both indicative of siloxane column bleed and tend to increase as the temperature increases; suggest that you disconnect the column from the MS and run a conditioning temperature program.