GPC of latex emulsion
22 posts Page 2 of 2
I can't just switch the solvent with progressive gradient solution (like starting with 90% THF go to 70% THF and so on until I completely switch to DMF)?
Unfortunately, PEA is correct. These polymeric packings swell and shrink during the conversion from one solvent to another. This is why there are columns available that are shipped in THF and others that are shipped in DMF. There is a real risk that the columns that you are converting yourself may not survive this procedure. Please check the C/U manual!
and if I switch to DMAC like PEA suggested? Is it the same problem?
Which solvents can I use with column shipped with THF because I don't want to change all the column bank...
The THF columns can be converted to HFIP or to more non-polar solvents like methylene chloride. I do not have all relevant information here, but the C/U manual that came with the columns should have this information, including conversion protocols.
However, all things considered, if your polymer is soluble in DMF, I definitely recommend to use the DMF columns with the DMF. You may have a hard time finding a more non-polar solvent that does as good a job as the DMF does. You are going into the wrong direction.
Thanks to all for your collaboration, you have answered all questions.
I'm back again with another question:
I need a second opinion on this: Could I run my THF-shipped columns with a mobile phase of 50% THF / 50% DMF because some of my latex are soluble in THF and some in DMF. I talked with a Waters technician and told me there's no problem as long as both solvents are miscible.
In total there are 23 users online :: 1 registered, 0 hidden and 22 guests (based on users active over the past 5 minutes)