Hi, all,
what does slope and intercept changing means for the calibration curve of a method? (r value is always better than 0.9995)
is a method is regarded validated to used if slope and intercept changes along time or fixed only between the instrument maintenance?
Thanks a lot!
emily lee
in my methods, did note a variation in slope if I ran the sets of standards many times over a 24 hour period.
Never understood the systematic drift.
Finally just made sure I met the criteria set for QC samples in the following document..
http://www.fda.gov/cder/guidance/4252fnl.pdf
seemed like high standard deviations to me, but did turn out to be a realistic goal..
Sailor