deuterated IS
Dear all
I'm trying to develop a LC-MS/MS method to calculate concentration of total homocysteine in plasma samples, but I'm in trouble!
well: I use deuterated IS: d8 homocystine (50µM in CH3OH with 1% HCl)and the reduction step has been perfomed with DTT (500 mM in H2O solution ) for 15' T ambient (Flow solvent H20/Acetonitrile)
the scan is MRM with their specific transition
after optimization of parameters of natural Hcy, I've begun to study the method with acqueous samples :I added in a vial natural homocysteine 50 µl (50 µM in H2O) + 50 µl d8 Homocystine (50 µM that will give, after reduction, the d4 Homocysteine 100µM)+50 µl DTT , after injection in MS/Ms system I've got a strange behaviour of the analytes :
the intensityes of D4 homocysteine is very,very lower respect of the natural homocysteine one !!!
insted the intensityes of the natural is very good
I thought IS intensityes must be more or less two time more!
Why??
the ionization behaviour of a deuterated is not the same (I mean proportional) of the the natural?There same problem with reduction-reoxidation , ionization or.......?
I need help to understand where is my mistake!!!
I've also change the reductor with TCEP but it caused seriuos ion suppression
thanks in advance to all of you
M.C.
I had samilar problem when I tried to use D8-compound as IS. My case is worse, I didn't see any d8 signal.(only a simple solution preparation, no reduction! I post this question some time earlier on this forum). Our coclusion is that the compound lost its D8, but We condn't know why. We finally give up without clear conclusion.
emily lee
I would guess that in your case the problem is in your reduction step. Have you checked the transition 277->140 to verify that you do not have any of the original D8 Homocystine? A high background could also indicate on column degredation.
Another possibility is that during the reduction you have exchange of H with D... I do not know where the D are positioned in the molecule and if this exchange is possible...