Question on SIL-10A syringe setup
Hi everyone,
Im setting up our SIL-10A (autosampler) with the syringe and wondering if anyone knows where the right (and bottom) tubing go to?
That is, do I leave the right/bottom hole open, or ? Are these going to waste/drain?
The syringe and SIL are connected and communicating properly so no issue there.
The right hand side port is connected to the needle inside the autosampler.
Bottom port , no connection. It is leakage port in case of rotor problem or overpressure.
Correction : the right hand side port is connected to washing solvent.
Other port is already connected to autosampler needle.
Thanks uzman for the reply and correction!
Can I use the same mobile phase (stored in a different bottle) as the 'washing solvent'?
Hello lespaul ,
Recommended washing solvent is iso-propanol but you may use other solvents which is best washing properties for the injected compounds.
In total there are 23 users online :: 1 registered, 0 hidden and 22 guests (based on users active over the past 5 minutes)
Users browsing this forum: Google [Bot] and 22 guests