Hydrogen versus helium as carrier gas with opium alkaloids
Client is seeking exemption from DEA/ICE regulation on natural products imported for decorative purposes. These are to be run on GC/FID because the client doesn't require high sensitivity.
I have attempted to replicate two methods found for running underivatized opium alkaloids on GC; still can't get good peaks with the standards. However, the methods I have found all use helium as carrier gas, while we run hydrogen. Is there any reason these alkaloids wouldn't show up using GC/FID using hydrogen as carrier gas? The other parameters (pressure, linear velocity, timing, etc.) have been adjusted for hydrogen from helium.
Example Restek chromatogram: http://www.restek.com/images/cgram/gc_ph00235.pdf
Hello
In my opinion it'd be better if you paste problematic chromatogram with full method description (parameters) rather than sample chromatogram from Restek website.
Regards
Tomasz Kubowicz
To get a chromatogram that good you would need a brand new column and a base-deactivated inlet liner. What you use as carrier gas will not make a difference.