retention gap and guard column
could someone please explain differences between retention gap and guard column?
So far I understand both of them are connected between the injector and the analytical column with same diameter as the as the analytical column but for different purpose..
guard column - to protect analytical column from contamination (e.g. non-volatiles in split injection mode)
retention gap - can act as guard column; does not protect the analytical column like guard columns do but used primarily for focusing the sample components when the sample is injected directly onto the column..
Regards
Ralph
Essentially they are the same thing and can serve dual purpose. If injecting with oven temperature below boiling point of solvent you will get a thin layer of solvent condensed on the bare column and it is considered a retention gap, but it will also trap high boiling and non-volatile components before they can reach the analytical column which makes it a guard column. Name mainly depends on the injection technique. You can inject 5ul in splitless into a cold guard column and it becomes a retention gap, instead of needing an on-column injector.
Hello
If you have thermally unstable analytes COC sometimes is the only option. S/SL inlet will not replace cool on column in all cases.
Regards
Tomasz Kubowicz
It has two purposes, but I think the guarding purpose is more than the gap. But, it fact it's a gap.
