Importance of Y-Intercept in Method Validation
5 posts Page 1 of 1
Dear Friends
What information we will get from Y-Intercept in method Validation
G.R.Reddy,
AR&D
BIOCON Limited,
India
Quick answer:
Intercept above zero: interferences, blank peaks
Intercept below zero: adsorption of substance
Thank you very much
Is there any articles or publications on this
There can be of course be other reasons for intercept not equal to zero (e.g not linear response). But if the respones is linear the explainations above are quite likely.
Regarding a reference, I think you can find it in a good LC book like Snyder-Kirklands "Introduction to modern liquid chromatography".
Also, you should compute the confidence interval around the intercept. If it includes zero, then you can consider the intercept to be zero for practical purposes.
Mark Tracy
Senior Chemist
Dionex Corp.
5 posts Page 1 of 1