Early eluted peak
I'm using 1200 Agilent HPLC with Max-RP column (2.00 x 50) 4 (µm).
mobile phase is 5mM Ammonium formate in 10% methanol and 5mM Ammonium formate in 90% methanol.
My analyte retention time is 0.5 min with 0.5 mL/min flow rate.
the dead volume is 0.1 mL.
I want the peak to elute little bit later than 0.5 min.
I tried to reduce the flow rate to 0.25 ml/min but it showed a border peak.
I'm thinking to increase the buffer concentration in my mobile phase, what other solutions do you suggest that might help me to get a later RT.
No way to make reasonable suggestions without more details:
- what kind of analyte?
- is your separation isocratic or gradient (you mentioned two different mobile phases)
- if isocratic, what is the actual composition
- if a gradient, what are initial and final composition and ramp rate?
In total there are 19 users online :: 0 registered, 0 hidden and 19 guests (based on users active over the past 5 minutes)