Negative y intercept LCMS
Hi all,
What are some possible causes of a negative y intercept (i.e. a positive x intercept) in a calibration curve (ESI LCMS method)?
The standards seem to have been prepared correctly (they're MeOH based standards spiked into a blank urine matrix).
Thanks for your help :)
Welcome to the forum.
This is a hardy perennial. Possible causes include non-linearity, adsorbtive (or other) losses of analyte, a signal threshold set too high, not taking the purity of the standard into account, poor repeatability, .....
What your particular problem is will depend on the details of your analysis - tell us those and we will be able to help.
A search of the archives will throw up plenty to keep you busy also.
Hi Peter, thanks for your reply.
So the issue is with Pholcodine. I use a mass-labelled codeine as internal standard.
I'll double check the purity when I'm in the lab next week.
For adsorbtive and other losses- do you think degradation of analyte in the standard master is likely? While codeine is reasonably stable in MeOH, I'm not sure about Pholcodine.
Thanks for your help.
Users browsing this forum: Google [Bot] and 5 guests