DSQ gain calibration failure
Page 1 of 1
I have problem on trace DSQ. (gain calibration failure). I have cleaned system. Ion volume has also been changed but without success.Does anybody know how to solve the problem and how to proceed?
It's possible that you have a leakage on your ion source heater. You can run diagnostics and look if you have a high leakage current. If so, you have to replace the ion source heater. Then tune and calibrate again. Keep me posted.
Michel
Michel
Page 1 of 1
In total there are 7 users online :: 0 registered, 0 hidden and 7 guests (based on users active over the past 5 minutes)
Users browsing this forum: No registered users and 7 guests