Recycling & Peak ReIntroduction in Semi-Prep. Chromatogr
Dear Chromatographers,
Is recycling & peak re-introduction doesn't goes well with semi-prep (upto 50ml/min) applications and hold good for prep chrmatography only??
Please post your views...Any theoritical justifications?
Rahul Yogi
rahul.yogi@gmail.com
Hi
For recycling, the system shold have some specs in order to do it well:
The pump should suck & pump constantly to avoid contamination of the recycled peak.
The control software should be able to handle it.
If you don´t need software control, you can buid it yourself with two three-way ball valves
:D
Best regards
Chris
There is a chapter in my book "HPLC Columns" that deals with this technique and related approaches (pages 267 ff).
In brief: recycling through the pump is not a good approach due to bandspreading in the pump. A much better approach to recycling uses a "walking column". In principle, you constantly switch the relative position of two columns. First you go from column 1 to column 2. Once all your peaks are in column 2, column 2 becomes column 1 and column 1 becomes column 2. Now you are only dealing with the bandspreading in switching valves, which can be made very small. This is a reasonable technique to get a lot of plates for a small group of tighly eluting bands.
It's easy to do if you have a six ports/two positions valve, in some softwares, you can set the switching program. I did it once, but at last gave it up, I couldn't bear the low efficiency, so I prefered to spend some time to find a better method.