CO2 free water for electrochemical detector
Hi all,
We are analyzing gentamicin by HPLC-ECD according to European Pharmacopoeia. And in the method they say use dioxide carbon free water for preparing mobile phase.
How could I prepare such the water. Is it OK if I use just DI water or bottled water from Merk.
Thanks for any suggestion.
The EP probably has directions to prepare CO2 free water. This is from the USP:
Water, Carbon Dioxide-Free—It is Purified Water that has been boiled vigorously for 5 min or more and allowed to cool while protected from absorption of carbon dioxide (CO2) from the atmosphere, or Purified Water that has a resistivity of NLT 18 Mohm-cm.
Thank you very much !
That said, I suspect that He sparging would get you there too. Once left exposed to air, your water will start to re-equilibrate to about 3ppm CO2.
DR
Electrochemical detection is sensitive to a lot of organic compounds. I therefore would not use DI water as it still can contain organics.
I recommend to use ultrpure water out of a purification system applying UV digestion (Millipore, Barnsted ELGA etc.).
Dr. Markus Laeubli
Manager Marketing Support IC
(retired)
Metrohm AG
9101 Herisau
Switzerland