< or > LOQ
17 posts Page 2 of 2
Are you trying to say what the value is or are you trying to show that the result is less than a value (at some defined degree of confidence)?
I'm trying to say tha value.
this is a classic EP method.
One way of thinking would be to make just one injection. Nowhere says that you have to inject twice the sample. EP is ok with that.
Another way is to take the worst case. If the average is above disregard limit, take the average. If it is below , take the measurment that is above disregard limit.
In total there are 18 users online :: 0 registered, 0 hidden and 18 guests (based on users active over the past 5 minutes)