SRI8610 GC for CO2 Detection, using mole sieve column
Hello All.
I am currently using SRI 8610 GC for H2 Detection, using
Mole sieve 5A 80/100
Argon as Carrier gas
Oven Temp- 70 C
Injection volume - 0.5 mL
Can I still detect CO2, with the same column, while still detecting hydrogen gas.
Please Suggest!!!
Ramana
No, not possible. Replace the ms5A with ms13x or preferably a 3m porous polymer column.
best wishes,
Rod,
If I replace MS5A with MS13x, will I be able to still detect H2,N2,O2; along with CO2???
Will detection level of hydrogen changes?
Ramana
It may take a longer MS13x than a 5A to get the same separation of O2 and N2 but yes you should be able to do the same analysis.
There is usually more tailing on N2 and O2 when using a 13X rather than a 5A but it depends.
Send me you email address and I can send you chromatograms.
best wishes,
Rodney George
Senior Research and Development Scientist
Gas Separations Research
Supelco
595 North Harrison Road
Bellefonte, PA 16823
814-359-5737 voice
814-359-5459 fax
rodney.george@sial.com
Rod,
So you are pretty sure, that MS13X will still have similar characteristics, for hydrogen separation, and hence hydrogen detection, using my system.
You can mail me at gadhamvr@nmsu.edu
Ramana
I hope you got the chromatograms of MS5A, MS13X, and Carboxen 1000 I sent you.
best wishes,
