split & splitless
4 posts Page 1 of 1
please give information of in split mode 1:100 that means how much sample taken
why we are useing RI detecter as a universal detecter,every compound in universal different RI .
why not find out all compounds by UV detecter.
I'm assuming that 1/100 of the flow is being used. So 1 mL/ 100mL.
As per my understanding every 100 parts of sample goes in to the GC 1 part goes to the injector and 99 parts purged.
a split of 1:100 means: from e.g. a total flow of 101 ml/min 1 ml of carrier gas goes to the colum and 100 ml go to the split vent.
Best regards!