Hi all,
I will try to adapt an HPLC-EC method for SAM (s-adenosylmethionine) and SAH (s-adenosylhomocystein). Virtually all of the literature data are based on colometric detection. I will only have access to a Recipe EC3000 with EC4000 cell (model sputnik). I don't have the manual yet.
I may handle the sample prep and separation but I don't have much experience with EC.
Are there great differences between amperometric and coulometric cells?
Can I run an amperometric cell and have a response comparable to coulometric cell?
Best regards,
Foren
Foren
Back from my job in instrument development I have the experience that there is almost no difference between amperometric and coulometric detection. In amperometric detection usually approx. 10 % of the component is converted (coulometric detection 100%). The lower signal of the amperometry is usually compensated by the higher background/noise of the coulometric detection.
Dr. Markus Laeubli
Manager Marketing Support IC
(retired)
Metrohm AG
9101 Herisau
Switzerland