Varian Pump 9012/9002 or 9010 fault 68
Hello, Does anyone have experience in diagnosing a fault 68 Pump Stall on a Varian 9010 or 9012 pump? The operators manual suggests to check hydraulics for clogged line. I am getting mobile phase flow through the column. The motor/piston make three turns and then stops with this fault. I cannot see anything hindering the movement of the motor and piston. I swapped out the optical sensor with a working unit and this did not help. The pressure goes no higher than 20 barr. Thanks!
Do you check plunger seal ? (or try to change new one plunger seal first)
Sometime system fault if system pressure do not meet requirement,
GiMiTEC | https://gimitec.com/catalog
Make sure that upper pressure limit of the pump adjusted more than 20 bar !
Users browsing this forum: Bing [Bot] and 5 guests