splitter advice
Well here is the problem I am using LCMS with ESI probe . The Mass system is waters ZQ4000 and what I want to ask is what type of splitter is best . eg low pressure , high pressure splitter valves. where i need only 0.2ml to flow to the mass . Will PEEK material or steel be better.
Also please suggest me good post column switching valve( Company).
Thank you
John
Hi John,
Why dont you use columns with smaller ID's, so you can work with lower flows?
Thank you
John
in my opinion,you should use a PEEK material splitter.because PEEK material can preferably resist your solvent and sample properties.
vicou
I have experimented with different setups for flow splitters for our Shimadzu LCMS 2010. I was nor very successful with homebrew solutions (i.e. a tee with different lengths of PEEK capillaries). The reason is, that the backpressure of the ESI changes, and so the split ratio was not constant over more than a few runs.
Finally I bought an ASI Quick-Split variable flow splitter (http://www.hplc-asi.com/pdf/FSPB2003np.pdf) that was very expensive but worked satisfactorily so far.
Regards, Hartmut
Users browsing this forum: Google [Bot] and 10 guests