Gradient Mystery
I have a problem with a gradient assay that I have not encountered before.
The last peak at the end of the gradient keeps increasing in peak area and height without changing retention time or symmetry.
A diluent blank injection showed an extremely good baseline with no carryover peaks.
The run contains 4 analytes. 2 of the analytes are before the gradient starts and one is at the beginning of the gradient; but, these show excellent retention time %RSD and peak symmetry.
Any advice would be appreciated.
Could you give more details eg nature of gradient, wavelength, type of compound, column and a typical problem chromatogram
No Tswett
That problem resolved itself mysteriously. After running about 7 injections of standard overnight. The peak heights returned to the typical level, and I was able to run a 15 hour analysis without 'this' problem recurring.
However this is the same method for which I just started a new post titled, "B Vitamins HPLC Analysis." All of the information you requested in this post is included the new post.
I do not think that the two problems are related, since this increasing peak height thing is a first out of many months of runs. However, I could be wrong.
Let me know if enough information is included in that post to help you visualize what is going on.