cis- and trans-cypermethrin
Hi everyone........
I´m looking for a column that will resolve cis- and trans-cypermethrin isomers........ any suggestions?
Marina Guti
We're analyzing cypermethrin on a HP5-MS column.
We find 4 peaks for cypermethrin, problably 4 different isomers.
It al depends on the GC-program. Our Gc-run is aproximatly 45 min.
We've also had luck with HP1, HP50 columns.
Hi mstolk and jh1.....
thanks for the input....
I have performed an intensive search on the subject and found that chiral separation of Alpha-Cypermethrin and found that:
Alpha-Cypermethrin can be separated on the BGB 172 column, but only the enantiomers of the cis diastereomers can be separated, giving a maximum of 6 out of 8 isomers........
So, HP-5 gives 4 peaks and BGB-172 (betacyclodextrine based) gives 6 isomers.... I´m still looking for a system that will separate all 8 isomers........
The method was published by Weiping Liu J. Agric. Food Chem. 2004, 52, 755-761
Marina Guti
In total there are 14 users online :: 0 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Users browsing this forum: No registered users and 14 guests