I am using Waters Empower Pro. I know how to create a Named Group, by selecting relevant individual peaks, but I am not getting calibration curve using Named Group as my Curve Reference. In review window I see amount and curve for my Named Group. On report I do not see it.
It will be a great help if anyone has used Named Group as Standard and Calibrated.
Thanks,
Shahu
In the report publisher, right click your table, table properties.
Tab page Peaks, use peak type group peaks.
This should do the trick
Bart
It did work. Thank you bart for your help.