Column and Method for CO2 solubility in organic solvent
I am a new phd student and am trying to set up a gc-tcd for some research. I have an HP 5890 series II GC with a Carboxen-1010 PLOT capillary column.
I am trying to determine the solubility of CO2 in acetonititrile and varying v/v percent solutions of acetonitrile/TFE.
Do I run the risk of ruining the column by doing liquid injections? Any suggestions on starting temperatures and ramp rates?
Jonathan,
Packed inlet or splitless/split? I don't think you will hurt the column but I am not sure if you will have enough sensitivity with a TCD. Also, you will have an issue of number of sequential runs you can tack together before getting popped by the solvent.
I am in Denver on occasion on travel, would be willing to stop by and discuss with you on my next trip up there. aicmm at flash dot net.
Best regards,
AICMM
Users browsing this forum: Bing [Bot] and 10 guests