potency of Esomeprazole
Basic questions from students; resources for projects and reports.
2 posts Page 1 of 1
I have one question. We are using Esomeprazole magnesium dihydrate as working standard for analysis of Esomeprazole. The potency of that standard is 94.23 on as is basis . So in calculation we should use Esomeprazole magnesium dihydrate or Esomeprazole magnesium after deducting the water.Can anyone help me in this ?
I would use the potency factor of 94.23 (% w/w ?) to calculate the corrected Esomeprazole magnesium concentration from your original weighing.
"As is" implies that the water content has been accounted for in the potency calculation.
If a purity factor was, for example 99.5% w/w (by HPLC/UV/other method) in the documentation, I would adjust as follows:
(Weight -(Weight x % water content)) x purity
Also, don't post the same question in multiple locations please :evil:
2 posts Page 1 of 1