ETP multipliers
4 posts Page 1 of 1
Hi
I use Saturn 2000 and wondered if anyone uses ETP (SGE) replacement multipliers. How do you get on with them?
Thanks in advance
WK
I'm Sorry I Haven't A Clue - Just A Minute - The Unbelievable Truth
We just installed an ETP multiplier in our Saturn 2000 mass spec... It tuned without any issues but I really can't comment just yet on sensitivity versus the old multiplier. The column I installed turned out to be high bleed and I'm getting ready to install another by the end of the week. I will let you know how the sensitivity looks. I'll do the 1pg/uL OFN injection and report back.
Thanks for the reply rubyfan
I must say, the ETP multiplier works great!! I recently performed a 1uL injection of 600 femtograms/uL dimethylmethyl phosphonate, a nerve agent surrogant, and saw decent detection by the Saturn 2000 in EI full scan mode. So evidently the multiplier is working well, considering it's a 10 year old system! I must say, we've always gotten significantly better limits of detection on our older Varian than our significantly newer Agilent GC-MS.
4 posts Page 1 of 1
In total there are 16 users online :: 0 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Users browsing this forum: No registered users and 16 guests