mass balance in stability study
I am doing a stability study of vitamin D3 injection. The Assay method for D3 quantitation is BP based UV method whereas impurities are identified using a validated HPLC method.
The stability profile of the product at different time points looks like this
Assay: 106.2%, 106.5%, 101.5%, 98.2%, 97.2%
% impurities: 0.12% , 0.14% , 0.19% , 0.25%, 1.17%
My query is that active is broken down about 9.0% from initial time point to last time point but I don't see 9.0% increase in impurities (but only 1.17%). Are these results acceptable even if we cannot find the mass balance? Is this significant increase in %impurity enough for presenting stability results?
Note: None of the impurities have RRF of more than 1.5.
Shahnaz.
May I ask what was the initial purity of the Raw Material that has been used to prepare stability batch ?
The initial purity of Raw material was 101.0%. Stability batches have 5% overage.
regards
Detection type, is the missing imppurities invisible to UV(or the wavelength) do toy have PDA/DAD to take 3d info. Can you run it on LC-MS?