GC analysis of ethyl difluoro acetate
3 posts Page 1 of 1
Hellow,
I am praveen , I want to know that How we can analyse ethyl difluoro acetate on GC , whether it's flash point is just 25°C. It is safe to analyse on GC or e should tried it on Normal phase HPLC . waht is right possibility please tell us.
Thanks and regards
Praveen
praveen,
I looked at the structure on the NIST web site but unfortunately they do not have any retention indices. I should think this compound would be doable via GC by either ECD or PID or FID. I should think a thick film PDMS at low temp would do the trick. Good luck.
Best regards.
Porapak Q packed
Flurocol packed
1% SP-1000 on Carbopack B packed
I would suggest a AT-steel tubing in lengths of 6 to 15 ft.
Lengths depend upon the isomers and the time and temperatures preferred.
Alternatives:
Thick film capillary SPB-2250, Supelcowax 10, or SPB-Octyl would give best results.
3 posts Page 1 of 1
In total there are 7 users online :: 0 registered, 0 hidden and 7 guests (based on users active over the past 5 minutes)