I am currently trying to add canabis (THC and its metabolite THC-COOH) into an existing method I have for drugs of abuse. All my compounds elute before 7.2 minutes on a C8 column. When I inject THC and THC-COOH in mrm mode they elute at approx 8.2 minutes each but peak shape is very broad and splitting. I inject samples of human urine 1/20 diluted with methanol:water (50:50)
My gradient is as follows:
min Flow rate ul/min A % B%
0 1000 95 5
1.5 1000 95 5
6.0 1000 5 95
8.0 1000 5 95
8.1 1000 0 100
10 1000 0 100
10.5 1000 95 5
12.5 1000 95 5
I would be grateful if you could give me advice on where to proceed without having to choose another column. The analysis is carried out on an applied biosystems QTRAP doing mrm to epi experiments in positive mode.
Mobile A=1900:100 (Water:methanol +25 mM ammonium acetate)
Mobile B = 1959:40:1( Methanol: 2-Propanol: Formic Acid)
What is the simpliest way to improve peak shape based on above?
Many thanks in advance